[PSUBS-MAILIST] Pressure Transducer

Jon Wallace via Personal_Submersibles personal_submersibles at psubs.org
Fri Jul 24 21:46:32 EDT 2020


 Turns out this issue was nothing more than miscalculating the 1ATM point of the sensor because it is an absolute gauge.  I got that sorted out and was able to do some real world testing from 100psi to 1psi and the sensor is spot on with the calculated values for the transducer, as well as two other analog gauges used for comparison.  Quite happy with it now.
Jon

    On Wednesday, July 22, 2020, 03:16:22 PM EDT, Sean T. Stevenson via Personal_Submersibles <personal_submersibles at psubs.org> wrote:  
 
 One obvious error source is that barometric pressures reported by weather sources correct the local pressure readings back to sea level (WGS 84 0'). You need to correct the reported pressure for your local altitude, or measure it directly using your own instrument.

Sean



  